N-Nitroso Desethyl Amiodarone in Amiodarone HCl Tablets by LC-MS/MS Metodu

I’m trying to do some method development work for N-Nitroso Desethyl Amiodarone. Because the limit is so low, I’m not getting any recovery when I spike the sample. I’ve tried many columns, mobile phases, and solvents. Has anyone else experienced the same issue, or can anyone make any suggestions?

When you say the limit is low, does that mean you are working with a relatively high sample concentration?

If so, matrix effects are very likely your main issue. What does your sample preparation look like? It sounds like the API or finished product is simply dissolved in solvent without any cleanup.

Which columns have you tried so far?
Do you have a UV detector installed, or at least a way to monitor the API retention time?

If the API and the impurity elute at the same time (or very close), you will inevitably run into problems.

You could try to improve the separation by optimizing the gradient and/or changing the column chemistry, but with so little information provided it is difficult to give specific advice.

Another option is MS optimization. If your MS sensitivity can be improved by, say, one order of magnitude, you could dilute the sample tenfold, which often helps significantly with matrix effects.

2 Likes

Hello, may I ask what type of sample you have? Is it an active pharmaceutical ingredient or a finished formulation? Is it an injectable solution?

We have both the finished product formulation and the injectable solution formulation. We are testing both, but we are not getting any recovery.

What I actually meant was that the 100% AI limit is 18 ng/day. That’s what I meant by the limit. Also, if we dilute the sample too much, the sample concentration we feed into the instrument becomes very low, and we can’t see a peak (around 0.006 ppb). The API and impurity peaks elute at very close times, but we’ve tried many gradients and columns. We’ve been working on it for about two months now. We haven’t achieved any results. We’re unable to overcome the matrix effect.

Agilent Technical Document

Maybe try this one out.
Desethyl Impurity and API are seperated even on a quick run.
Maybe with a different gradient and flow this could work for you.

Have you tried NP chromatography ? Given that those compounds (API and desethyl) elute quite early on this application note on a normal C18 column.

Also changing pH could have an huge influence on separation here. I know LCMS is a little bit resricted on puffer selection but it would try different pHs and evaluate those results.