Is Sensitivity really the biggest challenge or is it sample preparation?

NDSRI Analysis: Is Sensitivity Really the Biggest Challenge—or Is It Sample Preparation?

An interesting 2026 paper in BMC Chemistry reports an LC–TQ–MS/MS method for trace-level determination of two levofloxacin-associated NDSRIs: N-nitroso methyl piperazine (N-NMP) and N-nitroso desmethyl levofloxacin (N-NDLOX).

What caught my attention was not simply the low LOQs of 0.6 and 1.0 ng/mL, but the overall analytical strategy.

The authors achieved >98% extraction using a relatively simple 0.1% formic acid/acetonitrile system with 15-minute sonication, while tablet recoveries remained approximately 93–100%. Matrix evaluation also showed a solvent-to-matrix calibration slope ratio of 0.94, suggesting limited ion suppression/enhancement.

More importantly, the study connects:
:microscope: Analytical Target Profile / ICH Q14
:test_tube: LC–TQ–MS/MS method development
:pill: Finished-product matrix assessment
:bar_chart: (Q)-SAR and CPCA-based risk assessment
:recycling_symbol: Green/White Analytical Chemistry principles

This raises some interesting questions for those working on pharmaceutical nitrosamines:

  1. When NDSRI recovery is poor in finished dosage forms, should our first focus be MS sensitivity—or extraction chemistry and analyte–excipient interactions?
  2. Is comparison of solvent and matrix calibration slopes sufficient to establish matrix-effect control, or should isotope-labelled internal standards, matrix factors and/or post-column infusion become routine expectations for NDSRI LC–MS methods?
  3. How transferable is a successful extraction procedure from one formulation to another when NDSRIs can have substantially different physicochemical behaviour?
  4. As regulatory expectations evolve, should “method sustainability” become part of the Analytical Target Profile alongside sensitivity, specificity, accuracy and precision?

For me, the key takeaway is that the next generation of NDSRI methods should not be judged on LOQ alone. Recovery robustness, matrix understanding, toxicological relevance and routine laboratory practicality need to be considered together.

Interested to hear the experience of colleagues working on NDSRI method development, recovery problems and LC–MS/MS validation in finished pharmaceutical products.

2 Likes

1: Sensitivity is the biggest lever in my opinion. If your method is sensitive enough, you can dilute the sample further and make your life much easier. Once sensitivity is under control, you can start optimizing other aspects, such as the extraction solvent.

2: The way most labs assess accuracy is, in my view, more of a matrix effect control than a true extraction efficiency study. Because such small quantities cannot be weighed accurately, compromises have to be made. I typically spike a small volume and let the solvent evaporate, but this means that the analyte is mainly deposited on the surface of the matrix. As a result, many variables remain uncontrolled.

One approach could be to extract the same sample (provided the nitrosamine is present) using different solvents, or to perform multiple consecutive extractions with the same solvent. This would provide a much better understanding of the actual extraction efficiency.

Post-column infusion is also extremely important. It provides a direct picture of what is happening “under the hood” of the chromatogram. Normally, this information remains hidden.


My nitroso peak elutes at the black mark. Prior to that, the API elutes and reduces the infusion signal almost to zero. Afterwards, the signal returns to normal levels, but I still observe approximately 50% signal suppression and recovery. We eventually corrected that by using a suitable internal standard.
When you see something like this, you can be fairly certain that your LC-MS method will be highly sensitive to small changes, because the analyte is being measured in a “valley” between two larger peaks, resulting in signal suppression. In my case, I performed a DoE study to obtain a complete picture of the chromatographic system. Without post-column infusion, these effects would not have been visible.

3: Of course, there is no need to reinvent the wheel for every new analyte, and many established procedures work well for a broad range of compounds. However, I really dislike seeing the same 30-minute runs being applied to every formulation. In many cases, run times could be reduced significantly with relatively little effort.

4: Sustainability should be a focus for every analytical scientist. We should all be aware of safer alternatives to problematic solvents such as DCM and evaluate them during method development whenever possible. That said, there will always be cases where only certain solvents or approaches are suitable. Nevertheless, as mentioned above, I have seen many methods where run times could be shortened substantially, resulting in lower solvent consumption and higher laboratory efficiency.

5 Likes